Hi bourgarit
If you are using the Trace Elements DRS (i.e. not the 3D Trace Elements DRS) and you select NIST612 as your external standard (i.e. calibrant) then you will only get concentrations for elements with accepted values for NIST612.
If you switch to the 3D Trace Elements DRS, you can select more than one reference material.
There is more information about trace element calculations and the DRS here: https://iolite.xyz/support/webinars/webinar_recording_3dte.php, but if you have any other questions, please let us know.
Kind regards,
Bence